IMPORTANT: We are not currently connecting any new customers. If you previously linked during the BETA phase of the anesthetic monitor integration, that program has been discontinued. Please be aware that our support team can only assist with how-to questions. Unfortunately, we can no longer provide technical assistance for issues encountered aside from the steps documented here: Troubleshooting: What if the surgery monitor is not working properly? Additionally, we no longer provide technical support for the Surgery Admin panel issues.  Any other connection difficulties you will need to contact your IT services to resolve them.

__________________



 

 To set up the monitor correctly, please follow the steps below. 




Set correct IP address, Gateway, Subnet Mask and Host IP for the multi-parameter portable veterinary monitor - BM3VET Touch, (If it’s not already done). For this please select the settings icon on the monitor using the monitor's touchscreen.



 

NOTE: Alarm + Function can be used to lock/unlock the monitor's touchscreen, buttons, keys, and rotary wheel:

  • to lock: Press Alarm + Function keys at the same time until lock icon is displayed at the monitor's screen;

  • to unlock: Press Alarm + Function keys at the same time until the unlock icon is displayed at the monitor's screen


In the Setup window please select the Network option:                


Then adjust the values as per the instructions below: 


Network Option Description:

MAIN VER - Version on firmware;

CENTRAL - must be BL;

DHCP - must be OFF;

HOST IP - IP Address of the PC where the Surgery Admin application (Listener) installed;

DEVICE IP - IP Address of the BioNet BM3VET Touch Monitor;

SUBNET - Subnet mask of the network, by the default is 255.255.255.0;

GATEWAY - It is a start network device IP;

MAC ADDR - It is a media access control address, is a unique identifier.


NOTE: IP address must be different for each device within one network, ie:

HOST IP - PC/Listener IPv4: 192.168.033.100

DEVICE IP - Monitor IPv4: 192.168.033.101

Default Gateway: 192.168.033.001

 

For more details please also refer to the official manual: https://www.bionetus.com/bm3vet-touch-screen-veterinary-monitor/